(Sorry for the delay—harvest is in high gear.)
I am new to GC, so I am unaware of how chromatographers usually measure carryover. I first noticed it when I was running two identical 10-sample sets one after another--the second sample to run of each pair always showed greater peak area, amounting to an average of 0.2% v/v EtOH more than the first one. After increasing my dilution factor, I now see an average increase of 0.1% v/v EtOH in the second vial of a pair that is spaced 10 vials apart. Even if I cap some blank vials at home, or try to purge the vial with an inert gas, I can't avoid at least a very small EtOH ghost peak (smallest I have been able to get is ~2 or 3x noise), which leads me to believe that there are some good EtOH hiding places somewhere. But if I take that ghost peak on a freshly baked column to be "0%", and then run a blank after each sample in a run, I can see ghost peaks of between 0.03% and 0.1% v/v EtOH, perhaps depending on both the concentration of the preceding sample and the total number of samples that have been run so far during that run or since baking the column (I will do more extensive carryover testing after adjusting some settings).
We have a split/splitless inlet with a direct column connection from the transfer line, I believe. The autosampler is a Tekmar 7000, and the GC is an HP 5890 Series II.
The TTB (Alcohol and Tobacco Tax and Trade Bureau), uses AOAC Official Method 983.13 from the 17th ed. This is a very general (and antiquated--1988) direct-injection/packed column method, which of course, gives no information on quality of results whatsoever! I agree with rb6banjo’s suspicion that it is likely little/no better than what I am getting now.
rb6banjo, thank you so much for that numerical comparison! And yes, the high concentration has been the root cause of most of the problems in method development; I was wondering why I couldn't find any recommendations for working with high concentration samples besides “just dilute them enough”—I guess the reason is people generally choose a different method for high concentration samples!
The most common method in use today for determining alcohol in wine is ebulliometry (boiling point), which is sometimes said to have ±0.5% abv accuracy and fairly good precision (provided the atmospheric pressure doesn’t change between calibrating and running samples) but in reality, sugars and other solutes can be major interferents. Distillation/densitometry is ancient but not popular due to long analysis time and large sample requirement, but it can get ±0.1% abv accuracy (and excellent precision, too). GC-FID was the first modern approach developed for beverage alcohol determination, but, as you mention, NIR-based methods such as Anton-Paar’s Alcolyzer are growing rapidly in popularity due to their ease of use and purported quality of results (
http://www.anton-paar.com/Alcolyzer-Win ... ifications).
I like the fact that a GC is a more fundamental piece of equipment (for when we develop methods for other analytes), but it might be tempting when some used Alcolyzers start coming up on the market…
I will take a look at that full evaporation reference--thanks!
Thank you everyone for all your help!