wiclef » Tue Oct 29, 2013 8:45 am
There are lots of reasons why you might not see a signal from the detector. So you need to troubleshoot systmatically, step by step ?
First; do you have the correct detector channel selected in the software, and are all the other settings correct ?
If you select detector signal on the front panel of the GC and do an injection do you see the value in the readout changing ?. If not what value do you see ?
If you light the flame and hold a cold, shiny surface (say a shiny spanner) at the detector outlet, do you see condensation.
If you touch the detector body with the spanner do you see the detector signal or the chromatogram trace change ?
If you inject the gas from a cigarette lighter do you see a peak ?
Peter
Dear Peter,
I am working with the person who wanted your help.
We have even changed the detectors and check them so we are now sure that the detector
is working well because I have seen the condensation. Let me see if I can see if the chromatogram
can change when I am injecting the gas from a cigarette lighter. I was thinking that I have a problem with the column. I have to check the packed column.
Regards,
WICLEF KAGISHA