I'm hoping that someone can help me with this very perplexing issue. I am running an Agilent 7890 GC utilizing a Cool On-Column inlet. My column is a DB-5/ZB-5 30m x 0.25mm x 1.00um (yes, I know this is narrow for CoC but our test mixes give beautiful, reproducible and linear chromatography). Our sample is mPEG6 mesylate in ACN.
Problem:
It appears that the act of injecting my sample on the column creates active sites, as evidenced by the loss of my alcohol peaks. Injection of subsequent blanks causes the resolution and peak response to deteriorate until my peaks are nothing but ugly lumps in the baseline. Baking does not solve the issue and the column appears to be dead at that point. I've also tried rinsing the column with ACN and MeOH (both compatible with the column phase according to the manufacturer) and that only seems to make things worse.
Upon visual inspection after the samples have been injected, there are "particulates" within the column. The obvious conclusion is that there is some non-volatile component within my sample that is depositing onto the column. However, I have only injected the sample 8 times! This is with a brand new column. I am using a retention gap/guard column and there is little to no evidence of anything on the guard column. All of the junk is in the center of the column itself. At 1uL injections of a 1ug/mL solution, I cannot fathom that I would be able to see residual particulates with the naked eye. Especially in multiple places throughout the column.
My thought is that our sample is actually reacting with the column itself, however everyone I've spoken with disagrees. But honestly I can't imagine what else would cause such rapid destruction of the column. I have killed 3 columns in 2 weeks, one of which died within 24 hours.
We've run an analagous method for months using a split injection and have not observed this problem. My thought is that somehow injecting the liquid directly onto the column is causing the column to deteriorate much more rapidly than using a volatilized sample. But I just don't know. I've spoken to Phenomenex, Restek and Agilent and no one seems to have any clue what's going on.
I have added chromatograms to the link below. I can add any other images requested as well.
http://www.flickr.com/photos/104228113@N02/
Note that our Impurity Profile Solution is a test mix containing 29 impurities and intermediates in our process. The linearity chromatogram included contains a subset of those peaks that demonstrate the formation of active sites most clearly.
Thanks for your help!