Separating two peaks

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
hello,

I'm doing a limit test of dichloroacetic acid and monochloroacetic acid for lauryl dimethylamineacetic acid betaine

The problem is that the peak of dichloroacetic acid and a one of the peaks from the test substance have the same retention time.
which parameters should i change in order to separate those two peaks?

I hope the question is clear...

Thank you.
Use a column with a different stationary phase.

Peter
Peter Apps
You have given no conditions, so if you are using a capillarly column, set the flow such that the linear velocity in the GC column is between 22 and 35 cm/sec. If the flow is faster than this, column efficiency will be degraded. And, if you are looking at peaks that elute early in the chromatogram, sometimes dropping temperature or oven ramp rate can help.

Let us know what instrument, conditions, etc. you are using and show a picture of the chromatogram and help may be a bit mroe specific.
3 posts Page 1 of 1

Who is online

In total there are 57 users online :: 3 registered, 0 hidden and 54 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot], Google [Bot], Semrush [Bot] and 54 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry