Gaston » Tue Sep 03, 2013 2:18 pm
If your valves are in an external box, you're ok with oven temperature. The box is probably heated to a temperature warm enough to inhibit condensation of your analytes but not so hot as to melt your rotors. 300 °C detector temperature is plenty to get rid of fingerprints/oils, etc.
House air has a lot of junk in it. It's not designed for chromatographic applications. It's only good for actuating pneumatic valves, etc. Your baselines with get much more quiet when/if you switch to zero grade air. If you are not looking for "microsniffs of stuff" in your samples, your air might be fine. Just know that you won't have optimum detection limits with the house air. Also, the house air might fluctuate some as the compressors cycle and age.
Thanks for the tips. If I end up using the FID, I think I would have to change to better quality air, as I think that the amount of analytes in the sample will be rather low. But I hope that I will be able to do my analysis using the TCD only, hopefully!
I have noticed that the FID baseline has gotten a bit less "driftey" in the past days, though, and now it's mostly stabilized at 21-22 pA (chromatogram coming up).
The tee is probably fine for whatever temperature in the oven. No moving parts. The rotors in the valve have to be flexible to make good seals.
Seeing peaks when you inject?
I injected the contents of the sampling loop (i.e. air) through the inlet in splitless mode. In contrast to the chromatogram I posted earlier, I get a huge broad peak instead. The two "jolts" in the FID, which occur simultaneously with valve switching on and off, still remain. I also installed ChemStation MSD to control the GC and acquire data, seeing as this is the software I will most likely end up using. With the same settings in the two programs, the obtained chromatograms look, well, similar.
Using OpenLAB CDS ChemStation Edition
Using ChemStation MSD D.01.00
In ChemStation MSD, the y-axis is "response" instead of microvolts or picoamps. Is there a way to change the settings so that the "real" signal is displayed instead? (although I can imagine this will be rather irrelevant in the end for the integration and quantitation).
I feel like a bit of a monkey here hitting random keys in the GC, but from this very blunt method, I suppose it is fair to say that the equipment works mostly alright

I'd like to try something more methodical, though. As far as I know, this column is for separating gaseous hydrocarbons (GS-CarbonPLOT), but I think I don't have anything in the lab at the moment that could work as a suitable "problem sample". Any suggestions for that? Maybe inyecting a mixture of pentane and hexane?