*those such gradients are designed to selectively elute analytes that too hydrophillic to be separated using an iso-m.p.i would like to know why in some gradients,in the m.phase, there is a 5 to 15% isopropanol stable percentage(except the aquas and the ACN/MEOH(usually) portions whose percentage is changing).In addition i would like to ask if large amounts of isopropanol could damage a column.
thanks,
I've had columns come back with plastic "finger-tight" fittings used as "end plugs." Nothing surprises me anymoreFirst of all, what on *earth* would lead someone to store a reversed-phase column in hexane??!!
Hexane is not miscible with ACN/water. This is a situation where IPA can be very useful as a co-solvent: it is miscible with both hexane and with water. If you want to switch from a reversed-phase to a normal-phase system, an extensive rinse with IPA is the way to go.