We are doing trials with a new performance check that we developed in house from several references. Two of the tests are injector linearity injecting 5µl, 10µl, 25µl, 50µl, 75µl, 100µl, and another test is injecting 10µl of an solution at 50%, 80%, 100%, 120%, 150% concentration.
The limit we set from litterature was R²>0.998
but we are failing all checks if we plot results in force through 0 mode, and fail some of the tests if we plot without force through 0.
My questions are:
is R²>0.998 too tight?
for prooving linearity for performance test should the data be plotted in force through 0 mode?
The HPLC is set up at 1.5ml/min, 90% ACN, 40ºC. injecting a solution of about 100µg/ml Anthracene in 100% ACN.
BTW is there something less hazardous than Anthracene to use but which gives short runs and has a defined lambda max that can be used for wavelength calibration (as anthracene is Lambda max at 251nm)?
I know its a bunch of questions but maybe i'm won't be the only one to learn from this. Thanks in advance.