what is the appropriate time for SPME desporption time
Discussions about sample preparation: extraction, cleanup, derivatization, etc.
4 posts
Page 1 of 1
at 240oC, what is the appropriate desorption time for SPME in order to completely get rid of carry over.
-
- Posts: 960
- Joined: Thu Apr 18, 2013 2:10 pm
I've never really investigated that but that's the inlet temperature that I normally use.
I generally do manual injections so I base how long I leave the fiber in the injector on the cycle time of the GC. I leave the fiber in for 15 minutes while I'm equilibrating my next sample. Then I'm sure it's clean for the next sample. Most of the time, 30 minute SPME time (headspace is my main game) is good enough for my purposes. My typical GC run time is about 35 minutes.
If I leave the fiber in the inlet for 15 minutes + 30 minute SPME sampling is about 2 minutes longer than the GC cycle time. This maximizes my throughput for the day.
I generally do manual injections so I base how long I leave the fiber in the injector on the cycle time of the GC. I leave the fiber in for 15 minutes while I'm equilibrating my next sample. Then I'm sure it's clean for the next sample. Most of the time, 30 minute SPME time (headspace is my main game) is good enough for my purposes. My typical GC run time is about 35 minutes.
If I leave the fiber in the inlet for 15 minutes + 30 minute SPME sampling is about 2 minutes longer than the GC cycle time. This maximizes my throughput for the day.
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
For SPME I desorb the anaytes in the injector always 1 min at 260 on a carb/DVB/PDMS fiber then stick the fiber in either the needle heater or a different inlet for at least 10-20 min.
-
- Posts: 206
- Joined: Sun Oct 09, 2011 5:08 pm
Thanks a lot!
For SPME I desorb the anaytes in the injector always 1 min at 260 on a carb/DVB/PDMS fiber then stick the fiber in either the needle heater or a different inlet for at least 10-20 min.
4 posts
Page 1 of 1
Who is online
In total there are 13 users online :: 1 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 12 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 12 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science