BMU_VMW » Wed Aug 21, 2013 2:41 pm
Yes indeed: The loss of analytes happens every time I put in a new liner. It happens every time, no exception (but it took us a while to notice the connection between the liner and the problems with these analytes).
In the methode we are using we try to add new compounds on a regular basis. The analytes that are affected had been added and first we thought there was something wrong with sample prep, SRM, ... it was only later that we noticed the connection with the change of liners. Our 'older' compounds are not affected. In the worst case we had; about 10 compounds were seriously effected. There are 112 compounds in the methode at this time.
It doesn't always take the same amount of injections to clear the problem. (at least 20 at most 35).
We are a drinking water company and use this methode on DM-extracts of raw waters (surface and ground) and on drinking water. matrix is relatively clean.
- We do one run a week (60-80 injections which take +/-40 minutes/sample)
- We replace the liner every 150-200 injections (i.e. after 2-3 runs)
- septum is changed befor every run
- we change our pre-column every month and than the analitical column is trimmed (+/-30cm)
- 2-3 times a year the linerspacer is changed.
detector is a Thermo TSQ quantum triple quad ms.
since we had problems with these liners loosing some off their glass-beads and jamming the splitt-valve we have a small filter installed just before the split-valve. This filter is replaced and the housing is cleaned about 4 times a year or when a liner accidently breaks in the PTV. but this has no effect on the loss off analytes.
Hope this info can help.