How to know TCD failure?

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
I am using Yokogawa Series 1000 Mark 1 GC to analyze Paraxylene.
The Chromatogram that i got lately seems to be very weird.
The baseline drop from beginning for certain time, then it goes up then only peak appearing.

I had replaced all the columns and even TCD for the GC. All the carrier flow had been checked and doing good.
How to know a TCD is working fine? Any way to check?


Thank you.

Alice :)
Pray tell, were you, are you programing your oven temperature?

Test the TCD by doing an

Isothermal run

Inject air.

single peak with no tailing should be result.

best wishes,

Rod
i fix my oven temp to 80 degree Celcius (Factory setting).
Tried checking baseline, Baseline okay, no noise/interruption.

:?: :?:
maybe your injection system (syringe or valve) grab some air or water vapour which significantly change thermal conductivity of gas in your analytical channel.
Try to inject something inert at lower volume. Air, another gas (e.g. nitrogen in helium carrier).
You can check the resistance of detector
5 posts Page 1 of 1

Who is online

In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry