Septum purge shutdown

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
GC/MS 7890 has been shutting down due to inlet pressure. It seems unable to purge thus will not move onto next vial.
I have changed vent copper line and filter and it ran 20 vials then the same thing (liner/septa, septa nut all changed as well).
has anyone had a problem with the analogue valve malfunctioning so that it is physically unable to open valve? Any suggestions greatly appreciated.
Is the pressure too high or too low ? Have you checked for leaks ? - using a leak seeker not the onboard diagnostics.

Peter
Peter Apps
It is too high. Hovering at 32-33psi. Previously started run at 17.5.
No leaks found
A stuck back pressure regulator on the split is a distinct possibility. Check for flow out of the split vent. What do the pressure and the split flow do if the inlet is in splitless mode ?.

Peter
Peter Apps
4 posts Page 1 of 1

Who is online

In total there are 43 users online :: 1 registered, 0 hidden and 42 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Ahrefs [Bot] and 42 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry