varian MS 2000 Trap filament

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

6 posts Page 1 of 1
Hello everybody!

I have a problem with my Varian MS 2000.

I am getting a message "Trap filament is open" and am unable to operate the machine.

What should I do?
Please let me know.

This means one of your "two" filaments is burnt out. You can go into the software and initialize the other filament and you will be back in business. I can't remember where in the Saturn but one of the panels allows you the option of using filament #1 or #2. Switch to #2 and I bet you will be back in operation. Timothy

Thanks! vey much.

I did exactly what you said and it worked.

Thanks again!
I have the same problem right now and I dont know where to switch the filament 1 to filament 2, please help me!
Go to manual settings tab of MS work station software and on the 2nd filament if the 2nd filament is also open then you have to replace the filament
I get an error message "BOTH trap filaments are open. Please shut down the system and replace them."
Any help on that?
6 posts Page 1 of 1

Who is online

In total there are 179 users online :: 0 registered, 0 hidden and 179 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 179 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry