I'm testing ETO which is prepared in water near 20 ppm. I've been getting high RSDs since I developed the method (~3%). Now I'm getting 4-7% RSD.

I thought that I was not crimping tight enough on the headspace vial caps, but crimping tighter didn't solve the RSD problem.

Some notes on the method:

HS oven temp 90C loop 105C transfer line 110 inlet 200C

What can be done to improve thits? Could it be as simple as replacing a gold seal and liner?
MestizoJoe
Analytical Chemist and Adventurer
Venture Industries
Spider-Skull Island
I need to mention also that The RSD is fine for the less volatile sample which is ethylchlorhydrin (ECH). The RSDs for ECH have been consistent and always below 5%.

I thought the caps were not crimped tight enough since only the more volatile ETO has high RSD.
MestizoJoe
Analytical Chemist and Adventurer
Venture Industries
Spider-Skull Island
EtO is quite reactive. Are your numbers high or low?

I can see you need to have temperatures somewhat elevated for the ECH, but why the high inlet temperature? Would 120C be adequate? But then I am used to not using an inlet and I never had this issue. I never used a splitter and my results were always good when I left that variable out of the equation.

But I would replace the inlet and any other sites you might consider to be reactive.

best wishes,

Rod
Thanks for the reply Rod.

The numbers are correct if I take the average. I observe that after using my check standard. But they vary. I don't believe it decomposes but I may be wrong.

I'll change the liner and gold seal.

A newly realized variable that might be causing problems is the lab temp during sample prep. Been a hot few days here in the G and our HVAC system isn't too notch. I'm going to do sample prep in a better controlled environment from now on.

As for the temperatures I use in the method... When the method was developed my skills were novice. The conditions are not optimal. However during method development I saw good reproducibility.

Have a great weekend. And keep it classy Kansas City.
MestizoJoe
Analytical Chemist and Adventurer
Venture Industries
Spider-Skull Island
4 posts Page 1 of 1

Who is online

In total there are 24 users online :: 3 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Amazon [Bot], Bing [Bot], Semrush [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry