- Posts: 6
- Joined: Thu May 30, 2013 5:02 pm
could you please give me a SPE C18 extraction protocol?
Thanks
Advertisement
Discussions about sample preparation: extraction, cleanup, derivatization, etc.
thanks.
Could you please see whether this protocol makes senses?
I'm using SPE C18 (500 mg, 3 ml Column volumn).
1. Activate the column by using 2.5 ml 100% MeOH
2. Prewash the column with 8 ml 0.1% TFA pH 5.0
3. Load the sample (500 ml of serum + 1.5 ml of 0.1% TFA pH 5.0)
4. Wash the column with 2 ml of water
5. elute with 1.5 ml 25%MeOH in water
Can I increase the percentage of MeOH from 25% to 50%? Because I have to dry the samples after the elution step