Injecting solvent between samples?

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
I would like to know how common is to inject pure solvent (DCM, methanol, etc.) between sample (such as bio-oil, diesel) injections.

Some students (like those in my university) make a single solvent injection between every sample, but other operators says there's no need for that - a few minutes (5-10) near the maximum column temperature is enough for cleaning out any 'impurities' present in the column.

Any ideas?
I would guess that someone has been overflashing the injection liner and depositing residues in the pneumatics. Thus by injecting another slug of solvent they tend to wash out these residues so that don't interfere with the next sample.

If injections are done properly, there is no reason to inject a blank solvent between samples, unless you want to know your liner is dirty and absorbing residues or your injector system has been compromised.

My guess is things are not as 'pure as snow', thus the need for extra 'washings'.

best wishes,

Rod
2 posts Page 1 of 1

Who is online

In total there are 23 users online :: 0 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 23 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry