As a biologist re-formatted into an analytical chemist I feel your pain.
An Rx 5MS column would probably not be my first choice for this, but from what I can see of the chromatogram you are getting adequate separations in terms of retention time. The tailing on the peaks could be due to a deteriorated column, or to active sites in the inlet. If the latter a different column will not help, but it would be covered by routine inlet maintenance that your chemistry collaborator should be doing.
The advantage of using a liquid medium is that the liquid phase could be homogenized by stirring or shaking when you do the analysis. An homogenous liquid phase will give more repeatable partitioning of the analytes into the vial headspace, which will give more repeatable absorption into the SPME fiber, and more repeatable peak areas.
Are you taking repeat samples from cultures ? - if you need to be able to sample again from the same bugs, then you cannot use "destructive" techniques like heating the sample to get more volatiles into the headpsace, or adding large volumes of a dilute internal standard and homogenizing the sample into that.
If you are growing the bugs in sealed vials you might run into the issue of accumulating volatiles starting to inhibit growth, or particular metabolic pathways. Also, are these things anaerobic ?, do you have a controlled atmosphere in the vials ?
Back to the issue of internal standards and repeatability. Using an internal standard is very useful if you have steps in the analysis where volumes are variable - variability in dilution or concentration steps are perfectly compensated for by an IS, and the IS does not have to be a very close chemical match to the analyte. If the analysis includes partitions between two phases then ISs can still be useful, but then they must be as close a match as possible to the analytes - which is why people use isotope labelled internal standards. SPME is a partition-based preparation technique, which is why your ISs would have to match the analytes, and with the diversity of analytes that you have getting ISs to match all of them is going to be difficult.
You need to know what the source of the variability is, and whether it is the kind that can be fixed by internal standards. From the look of the chromatogram the tailing is probably compromising the integration, and adding an internal standard will do nothing to help this. In fact it will make the repeatability worse because the poor integration of the analyte and the IS peaks will add to one another. So you need to present your chemical collaborator with replicate samples that you know are the same - to produce these you need your analytes dissolved in agar at concentration similar to what you get with the bugs growing on it. Make up 5 vials with this spiked agar in it and no bugs. Take a SPME sample from each and get them analysed and see how repeatable the areas are. If they are poor, make up a solution of the analytes in a solvent that can be injected to the GC and run five replicates of that. If the repeatability is still poor the problem lies in the instrument, not with the SPME sampling, and adding an internal standard to your samples is probably not going to do any good at all.
You probably already know that there is a lot of literature in the forensics area about volatiles from carcasses and cadavers - I can send you some papers if you ping me a peter-j-apps-at-g-mail-com. Take out the dashes and make at an ampersand , it keeps the bots at bay.
Peter