I am a freshman in GC and my experience is unsufficient to fix the problem I met. We analyze beer for vicynal diketones - diacetyl and 2,3-pentandione. HS, ECD and a packed column (Mss316 Chromosorb101) are used in the analysis. Calibration curves for the compounds are good. But when I run a beer sample, the retention times of VDK increase significantly, e.g. from 3.85 min for diacetyl in the standard solution to 4.12 min for it in the beer sample. When I added the standard solution to beer, incubated it for a couple of days, nothing changed. Two peaks with the maxima at 3.85 and 4.12 min were observed. I checked also if the ethanol content of a sample affects the retention time. Again nothing changed in the retention times. I supposed the yeast biochemistry could affect the results of the beer analysis. Beer from another brewery was analyzed, again the retention times were longer, however a little bit less (4.09min), compared to those for standard solutions. So, I can not say that the problem results from contamination of the column, the effect of some beer constituent or yeast biochemistry. If somebody has any idea what may be the reason for this difference in the retention times of VDK in standard solutions and beer samples, please, let me know.
Thank you,
Alex