pressure drop during the run

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

10 posts Page 1 of 1
Hi,
I have a problem of our Agilent 1200 HPLC. The pressure is stable when I don't inject the sample, but after injection (around 5-6min) the pressure suddenly drops to almost 0 and increases again soon. Does any one know why is that?
thanks
xiumin
Post your gradient conditions.

If isocratic, you've really got issues !!!
It is isocratic. And there is a leak from the bottom of active inlet valve. I changed pump seal and inlet valve cartridge.It still leaks.
Should I change the Active Inlet Valve too?
It is isocratic. And there is a leak from the bottom of active inlet valve. I changed pump seal and inlet valve cartridge.It still leaks.
Yes, active inlet valves can leak over time; you state "leak from the bottom of active inlet valve".

Also, sometimes the two pistons in the pump head get scratched and need replacing. I'll assume that you actually changed both piston seals. Also, the outlet valve can leak and trickle down. And did you use a new gold seal when doing that active inlet valve cartridge?

If your leak is truly "from the bottom of active inlet valve", replace that. It "ain't gonna" heal iself.
thank you very much. I will change the active inlet valve. When you say gold seal, do you mean the gold seal in the purge valve or in the outlet ball valve? I change the purge valve gold seal but haven't change the outlet ball valve gold seal.
Can anyone tell me why the pressure drop only happens when I inject the sample? I just did the leak test, every thing is fine.
Check and replace rotor seal.
we just changed rotor seal and I ran the sample on the following day, the leak came out from active inlet valve again. So, I will change active inlet valve and see whether it will solve the problem.
Hi, the problem solved after I change active inlet valve.
10 posts Page 1 of 1

Who is online

In total there are 12 users online :: 2 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Google [Bot] and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry