Hi, experts:

I am a new student in the field of HPLC. I have learned several types of HPLC techniques such as reverse, normal, ion exchange, size exclusion HPLC and I understand the principles behind all of them.

However, I have some questions regarding the choice of solvent and the use of buffer.

1. Depending the pKa of the functional group, how do you choose the buffer and what is the best buffer range to work with?

2. When a buffer is said to have 30mM, does it mean milli molarity? Does the buffer mean mobile phase in HPLC? Also, how do you decide the concentration of the buffer in the mobile phase?

3. Is the use of buffer always necessary?

4. If my injection sample is prepared at pH5.0, what shall be my buffer range be prepared?

Sorry, the questions seem a bit too grade 10, but I will really appreciate any information to help me understand more about HPLC.

Thank You.

1. Retention can sometimes change dramatically with the degree of ionization of the analytes, and this changes most rapidly when pH = pKa. Because pH is a logarithmic function, +/- 1 pH unit changes the %ionization from 10% to 90%.; +/- 2 pH units changes the %ionization from 1% to 99%. Therefore, the best general recommendation for initial work is to stay at least 1.5 pH units away from the pKa of your analyte(s) if possible.

1a. In reversed-phase chromatography, hydrophobic ("non-polar", "greasy", . . . take your pick) molecules are more strongly retained than hydrophilic molecules. Given a choice between the ionized and neutral forms of the molecule, the neutral form (free acid or free base) will be more hydrophobic and therefore more strongly retained than the ionized form (anion or cation). Therefore, it is usually best to start with a pH at which the neutral form predominates (relatively low for acids; relatively high for bases). In ion-exchange, neutral species are unretained, so you would go the opposite direction on pH.

1b. Once you have decided on a pH, you need to find a good buffer at that pH. Remember that buffers are themselves weak acids or bases; they are most effective at their own pKa, and become ineffective more than 1 pH unit or so away from their pKa.

1c. Once you have a list of candidate buffers based on their pKa, you can weed out choices based on detector compatibility. Many organic buffers, for example, have fairly high UV cutoffs. If you are chromatographing proteins and detecting at 280 nm, this is no problem. If you have to detect at 210, then it is a problem. If you will be doing mass spec, then the buffer must be volatile. Most of this aspect is simply applied common sense.

2a. Yes, mM = millimolar. Be careful, however, because there is a bit of ambiguity here. I have always interpreted the buffer concentration and pH as being defined with respect to the aqueous part of the mobile phase (i.e., before any organic solvent is added). A substantial number of people interpret concentration and pH as being defined with respect to the final mobile phase (i.e., after addition of organic). Hopefully, the interpretation will be clear from context (e.g. "50% acetonitrile + 50% of a 25 mM, pH 2.5 potassium phosphate buffer" is different from "25mM, pH 2.5 potassium phosphate buffer in 50% acetonitrile/water"). If it's not clear, you may have to find the original method and look at the detailed procedure. Which brings up another point: if you are documenting a procedure, be as explicit as possibe in describing the mobile phase preparation!

2b. The buffer concentration needs to be high enough to prevent pH changes caused by the presence of the sample. This is usually in the range of 10 to 50 mM (in the aqueous part of the mobile phase!). I'd say the most common values are 25 mM for UV detection and 10 mM for mass spec. You could often get by with less, and you may occasionally need more, but most people don't attempt any optimization of the buffer concentration unless they have problems.

3. If your sample contains only neutral (non-ionizable) compounds, then a buffer is usually not necessary in reversed-phase. Otherwise, use of a buffer is highly recommended.

4. Buffer pH is usually selected on the basis of separation chemistry (see responses 1, above). One good rule for HPLC is "to get the best results, disturb the equilibrium of your system as little as possible". This suggests that the sample should ideally be dissolved in your mobile phase. If the sample must be dissolved in something other than the mobile phase (a buffer at different pH, in your example), you should not have any problems so long as:
- the injection volume is low and/or
- the pH difference is not too great and/or
- you have sufficient buffer capacity in your mobile phase to deal with the upset.

Finally, don't apologize. Basic questions are what this section is all about. We all had to learn either by asking questions or making mistakes (the former is a lot easier and cheaper!). :D
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Cokarette,

One additional bit of information regarding your third question: "Is the use of the buffer always necessary?"

In ion exchange separations, buffers are often used as Tom indicated to control the extent of ionization of the analyte or the stationary phase or both in order to optimize retention or selectivity. But it is fairly common to operate without any buffer provided the mobile phase guarantees that the analytes are 100% ionized. For example, when using a SAX phase with sodium hydroxide eluent, it's generally not necessary to have any buffer present. The mobile phase pH will result in most analytes being 100% ionized and under these conditions no buffer is needed.
3 posts Page 1 of 1

Who is online

In total there are 206 users online :: 0 registered, 0 hidden and 206 guests (based on users active over the past 5 minutes)
Most users ever online was 11068 on Sat Dec 06, 2025 9:55 pm

Users browsing this forum: No registered users and 206 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry