GCFID NOISY BASELINE
Discussions about GC and other "gas phase" separation techniques.
5 posts
Page 1 of 1
Our GCFID has got very noisy baseline all of a sudden. Do i need to clean the detector and how do i do that?
-
- Posts: 212
- Joined: Fri Dec 05, 2008 11:03 am
the baseline noise is cyclic. There is a peak followed by by a negative and this continues every couple of minutes
-
- Posts: 212
- Joined: Fri Dec 05, 2008 11:03 am
We recently changed our method to a pulse split method:
Injection Pulse Pressure 30psi until 0.5min
split ratio 10:1 30ml/min
The method itself is 2 min hold time at 40 degrees. Then ramp up to 170 degrees in 3.25 minutes. then ramp up to 300 from 170 in 4.5 minutes. Then hold at 310 degrees for 15 minutes.
The GCFID appears to stop intermittently after 100 injections then sometimes after 40 injections. Is the new method causing this? We are using N2 as make up gas. H2 and air as carrier gas
Injection Pulse Pressure 30psi until 0.5min
split ratio 10:1 30ml/min
The method itself is 2 min hold time at 40 degrees. Then ramp up to 170 degrees in 3.25 minutes. then ramp up to 300 from 170 in 4.5 minutes. Then hold at 310 degrees for 15 minutes.
The GCFID appears to stop intermittently after 100 injections then sometimes after 40 injections. Is the new method causing this? We are using N2 as make up gas. H2 and air as carrier gas
-
- Posts: 327
- Joined: Thu Jan 20, 2005 12:55 pm
I have a special heavy handled screwdriver that sometimes works on noisy FID baselines. They can be prone to dust building up and a couple couple of thumps with the special tool can fix the problem.
GCguy
GCguy
GCguy
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
If it is cyclic in nature I would check that all the electrical connections are tight.
I had a similar problem once and it was due to a freezer sharing the curcuit. Baseline disturbances when the freezer compressor cycled!
I had a similar problem once and it was due to a freezer sharing the curcuit. Baseline disturbances when the freezer compressor cycled!
5 posts
Page 1 of 1
Who is online
In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Baidu [Spider] and 24 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Baidu [Spider] and 24 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science