yeah,105 degree exactly。Are you headspace oven is at 105? If water is in the sysytem it should be maximum 85
Advertisement
Discussions about GC and other "gas phase" separation techniques.
yeah,105 degree exactly。Are you headspace oven is at 105? If water is in the sysytem it should be maximum 85
105 C is above the boiling point of water at 1 atmosphere pressure. Presuming that your vials are leak tight the water will not boil, until the needle pierces the septum and the headspacer valve switches to fill the loop by venting to atmosphere. Then the water will suddenly boil - spraying droplets into the gas stream that will end up in the loop and other parts of the sampler plumbing. This cannot be good.in all, usp467 method validation work,for me,not easy work,never easy-going。 i have to deal with deviations of validation work,upset。
why 105 degree not suitable for water?how to persuade the other?water in vial can't boil in 105 degree,although reach boiling point。
105 C is above the boiling point of water at 1 atmosphere pressure. Presuming that your vials are leak tight the water will not boil, until the needle pierces the septum and the headspacer valve switches to fill the loop by venting to atmosphere. Then the water will suddenly boil - spraying droplets into the gas stream that will end up in the loop and other parts of the sampler plumbing. This cannot be good.in all, usp467 method validation work,for me,not easy work,never easy-going。 i have to deal with deviations of validation work,upset。
why 105 degree not suitable for water?how to persuade the other?water in vial can't boil in 105 degree,although reach boiling point。
Peter
i am doing residual solvents GC method validation. encounter puzzle。
first prepare 0.1ppm benzene solution。then transfer into vial like this:
5ml water to 20ml vial,then transfer 1ml benzene solutioninto the vial using pipette,last cap. parallel six vials. then run sequence to analysis using agilent 7890-7697 GC system. the peak area data as follows:
13.9/13.7/6.4/6/5.7/5.6
bad reproductivity, what's the reason? not once, the phenomenon often emerges. usually the first vial peak area maximum,the other gradually decrease in area。