I'm using a 6890N Agilent GC column is 30 m 0.32mm, 0.5µ thin film of a 5% Phenyl-95% methyl silicone bonded phase.
Detector is a FID and injection is 1µl of air (empty vial). Condiitons are Carrier He 1.5 ml/min constant flow, H2 40 ml/min, air 400 ml/min, make up 48.5 ml/min of He. Split 1:100.
Temperatures Inlet 100°C, Detector 280°C and column 50°C during 30 min, then 20°C/min to 250°C, wait 30 min.
I'm finding the strange peaks you can see in the image. They are not column contamination because their shape is independent of the temperature.
Perhaps someone of you can help me to fix the issue.
Thank you so much :
http://farm9.staticflickr.com/8236/8502 ... 284a_b.jpg