We just recently had our 5890 repaired and placed back in service. We will be using it mainly for alcohols and glycols. What is the best column to use for this application?
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
624 or 1301 phase, or sometimes a Wax.
Rod
Rod
-
- Posts: 69
- Joined: Mon Jul 11, 2011 3:15 pm
If injecting water directly, wax type columns are the only way to go due to polarity.
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
This.624 or 1301 phase, or sometimes a Wax.
Rod
If injecting a water solution, keep the injection volume 0.5 ul or less.
-
- Posts: 35
- Joined: Wed Nov 28, 2012 3:46 pm
Yup. Gotta go with wax. Here's one: Restek Stabilwax Cat#10655. I use it for glycols with a drilled uniliner. Works great.
-
- Posts: 4
- Joined: Wed Oct 31, 2012 2:41 pm
Thank you everyone for your advice.
6 posts
Page 1 of 1
Who is online
In total there are 34 users online :: 2 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Google [Bot], Semrush [Bot] and 32 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Google [Bot], Semrush [Bot] and 32 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science