lmh » Tue Jan 15, 2013 3:09 pm
The first option is to look for a literature method with LC-electrospray and use whatever solvents the authors used.
The other option is to choose your own, in which case I would start just by using something simple like 0.1% formic acid. It may look odd to use an acid when you're hoping to see an acidic compound in electrospray, because the acid should, theoretically, mean that the acidic analyte is less ionised in solution, but electrospray is actually quite good at seeing things at pH values at which they "shouldn't" be ionised (for betulinic acid you'll have to decide whether to look in negative or positive mode. It may work in both). The main point of the acid is to give you a solvent pH which isn't very close to a pKa of any compound in the mix, for the benefit of chromatography. Since you'll be doing reverse phase, probably on a silica-based column, you're doomed to acidic pH values. If you were to need a buffer, you can use ammonium formate and ammonium acetate buffers, but I wouldn't bother unless it's actually necessary/beneficial to chromatography.
Routinely, when presented with phosphate buffer methods for HPLC with a need to run in LC-MS, I just replace buffer with 0.1% formic acid and see what happens. Depressingly, 90% of the time, it works.