For the most part I have two actives in methanol to which a DBP internal standard is added. The issue I am having is one of the analytes is giving consistently varied results even from the same sample. The other analyte is fine however. Also RRFs are reasonable..
I have found that some of the troubled analyte is being hung up pre-coloumn as a flush after the standard shows. The amount that is being hung up is varied as well. The pre-coloumn is tapered with glass wool packing which, when changed stops this 'sticking' of the problem analyte for a short while before the problem returns and slowly grows over time.
To add to this, when I do this whole procedure with acetone instead of methanol the problem is gone. This is with a split-less injection of which there is far too little analytes in the standard or sample to overload the coloumn.
Any ideas on what is causing this issue?