Make-up gas flow rate on HP 5890

Discussions about GC and other "gas phase" separation techniques.

7 posts Page 1 of 1
I am pluming an old HP 5890 ( not 5890A, not 5890 series II) with FID.

Use helium as carriers gas, while nitrogen as make-up gas.

I am checking the flow rate of gas supplies without running the GC/detector. The problem is the nitrogen makeup gas does not have any flowrate :cry: when I plugged in the bubble flowmeter into the FID outlet.

I think it is abnormal. I did not find anything in the GC manual to say one way or another, but I suspect that the control valve for the carrier gas is broken.

Anybody has experience on similar problems?
Just to be sure - you have plugged the bottom of the detector where the column would connect ?

The 5890 is a real vintage instrument - if I recall from thirty years ago there is a pressure regulator and gauge behind a flap on the left hand side that you use to adjust the makeup flow and a needle valve on the front to turn it on and off. If this is what you have - do you see pressure on the gauge ?

Peter
Peter Apps
Thank you for your reply.

There is no pressure regulator inside the left panel. I am hooking up the gas line directly to the detector gas manifold (whatever it is called, the gas connector with needle valves).

I found the problem though. The instructions are on page 11-4 to 11-5 of the manual: Setting capillary make-up flow rate. The revealing thing is that this valve for AUX Gas is not a typical valve, like the other two on the same panel. Instead a screw at the center of the valve needs to be turned. I suspect that many people might have missed this "NICE" design since the valve handle on this machine was kind of broken. I can image that people were trying to get more flow by cranking the valve like crazy :!: I now know for sure a person before me was not using the makeup gas because this restrictor was turned all way to the right = no flow.
I think it depends on the 5890 whether it is a 5890A series II or series IIa. My series IIa has no guages though I am thinking of buying Restek's guage pack. There is just a knob on the front pannel. There is also a black inline regulator on each of the three gas lines I believe.

I think the makeup flow to the FID is governed by pressure and a restictor frit similar to the auxilary EPC on my 6890. I haven't had to mess with mine so I'm not sure but there is a page in the manual about it.

Restek sells a nifty guage pack that mounts to the side of the 5890 for air and hydrogen but it doesn't inlcude makup.
https://www.chromspec.com/pdf/e/rk125.pdf
part 22071
Are you using a packed or capillary column?
Where can I buy the kit they use in CSI?
I am using a capillary column.

Thank your all for your replies. As I pointed out in my 2nd post, I solved the problem.

As I mentioned there, there is a screw in the middle of the valve handle. This screw (/restrictor) restricted all the makeup gas flow. After finding the instructions in the 5890 GC manual, I was able to adjust the screw and get the flow rate I wanted.
Congrats on the long lost skill of reading manuals - its amazing what you can find in there. :D

I just finished showing some of our techs how to work that old flow controller - they were ready to run when I took out the bubble meter.
7 posts Page 1 of 1

Who is online

In total there are 30 users online :: 2 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot], WillieGon and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry