It appears that I misunderstood you. I supposed you were recommending that a good DMF measurement from the solid was a possibility. I missed the dissolve in NaOH part. Sorry about that.
Yes. We agree.
For what it is worth: I once developed a validated method to measure a volatile nitrogen base, present as a salt, in a solid matrix, a peptide in fact, using a total evaporative technique. It wasn't perfect, but was accurate enough to determine a minimum of 50 ppm reliably and accurately. This was adequate for regulatory needs at the time. And at $100 a milligram, using less sample was better. I was hoping for my usual 1 ppm, but sometimes the chemistry is what it is.
I also have done residual solvent analysis on nearly insoluble zinc salts by heating them to their melting temperature.
But I have never gotten reproducible results trying to extract solvents from solids, although how heartily I have tried. (Did get close once, but no cigar.)
Using NaOH (aq), temperature performs wonders, and sometimes a lot less than expected will get the job done.
DMF still is a trouble maker, as it loves to stick wherever it wants which is mostly everywhere. Temperature is a great help here too.
best wishes,
Rod