It is possible.  What I have seen in the past is that the filaments usually burn out before you notice a large drop in performance.  I wouldn't use the Filament #2 position for any analysis.  I've been told by Agilent reps that the performance of a filament in that position tends to be bad.  It's there more as a sales gimmick than anything else.  I would take out the source and swap the #1 and #2 filaments around and see if you get better performance.
Agilent said that the #2 filament tends to "sag" down away from the ion volume when it heats up reducing it's ionization effect.  Not sure if it's true because I always use filament #1 to be safe.   Anyway, these filaments should last you about 1-2 years running daily if you are not blasting them with solvents.
P.S.  Nice screen name.  If you're in a contract environmental lab it doesn't get much better.
Ty
That bit about the filament #2 "sag" is very pertinent. I've installed my two 
SISalloy plus Yttria coating in my 5973 and the first calibration attempt was a disaster. It seems that since the new filiments are not as bulky as the old style filaments so I decided to position the filaments as far from the source body hole as possible. This increases the gap maybe 0.1-0.2 mm between the source body and the filament.
I use a 70 ppb ISS and 20,40,80,120,160,240,320,400,480 ppb standard curve. What happened is that response for the second half of the chromatogram (higher boiling point) compounds rolled off dramatically at about 160 ppb. I tried again the next day with the same result. So, being in a time bind for getting some samples done. I did the following.
I vented and replaced the SIS filament with my remaining original filament in the #1 position and left the other SIS filament in the #2 position. I made sure they were snugged in the opposite direction. So filaments as close to the source body hole as possible; and of course aligned parallel to the hole. I figured I would try to use the SIS filament in #2 first and if that failed, fall back on the old style unused filament in #1.
I went for vacuum and the next day put on a full calibration using the SIS Filament #2 and the Atune hat trick; I went for a 219/69 ratio of 55%. Ramping the repeller gave me a nice set of responses for 69, 219, 502 centered on 30 volts. Its never shown so "clean" before with the standard filaments. Maybe that was just luck with such a good vacuum after only 12 hours and having opened it three times after cleaning: water 0.90%, Nitrogen 1.18%, Oxygen 0.5%, Carbon Dioxide 0.36%. I purged a 240 ppb standard and saw that it had very similar response to my old calibration back in May 2017. I went ahead and put on a full calibration.
So, by 10 pm I had a very nice calibration. Most compounds had an RSD of 4-7%. My dibromofluoromethane, 1,2-dichloroethane, toluene-d8, and BFB were respectively 1.04%, 1.58%, 0.63%, and 0.85% RSD. I am using helium carrier gas. Compounds that have been non linear in the past or were affected (low response due to decomposition) by hydrogen carrier like carbon tetrachloride 2.01%RSD, dibromochloromethane 2.91%RSD, bromoform 4.55%RSD, and 1,2-dibromo-3-chloro propane 5.63%RDS were good from 20 all the way to 480 ppb.
I had a little higher RSD for acetone 12.3%, diethyl ether 10.4%, MEK 14.3%, tetrahydrofuran 14.6%, and MIBK 11.7%. But, I think that is more of a purge and trap issue. I'm purging at 70C but still using a water management coil rather than a bypass line on my Tekmar 3000.