-
- Posts: 10
- Joined: Fri Nov 22, 2019 8:52 pm
Polyamide oligomers by FFF - MALS
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
2 posts
Page 1 of 1
Does anyone have any experience in analysis short (up to 5000 Da) linear oligomers of polyamide by FFF - MALS? Currently my equipment (in aqueous mode) allows me run analysis with eluents only up to 20% of methanol. Is it even worth to try analyse this oligomers in such a mode?
-
- Posts: 23
- Joined: Thu Feb 21, 2019 11:18 pm
Hi,
While it is possible, FFF-MALS is not really a good solution for such small polymers, up to 5000 Da. Current FFF-MALS instrumentation from Wyatt Technology can certainly handle higher concentrations of methanol, and you can use regenerated cellulose ultrafiltration membranes though you might have to cut out your own membranes due to the low MWCO required (the lowest standard size is 5 kDa but you would need 1 kDa or less, depending on the lowest size present).
I recommend SEC-MALS for this analysis. Then you will not have to deal with finding low MWCO membranes or any solvent limitations.
best,
Dan
While it is possible, FFF-MALS is not really a good solution for such small polymers, up to 5000 Da. Current FFF-MALS instrumentation from Wyatt Technology can certainly handle higher concentrations of methanol, and you can use regenerated cellulose ultrafiltration membranes though you might have to cut out your own membranes due to the low MWCO required (the lowest standard size is 5 kDa but you would need 1 kDa or less, depending on the lowest size present).
I recommend SEC-MALS for this analysis. Then you will not have to deal with finding low MWCO membranes or any solvent limitations.
best,
Dan
2 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: John Guajardo and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: John Guajardo and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science