If I am analyzing a protein-DNA complex by SEC, should I use protein calibration standards or DNA calibration standards?
-
- Posts: 23
- Joined: Thu Feb 21, 2019 11:18 pm
With SEC-MALS, you don’t need to use any molar mass calibration standards! The MALS detector directly measures the molar mass of the protein, DNA, or other macromolecule. The measurement is independent of the elution time and specific calibration standards. Whether your molecule has a different density or conformation, compared to your standard, and whether your molecule interacts with the column (thus influencing its elution time) does not matter for SEC-MALS.
[answer provided by Sophia Kenrick, Ph.D., Wyatt Technology]
[answer provided by Sophia Kenrick, Ph.D., Wyatt Technology]
-
- Posts: 23
- Joined: Thu Feb 21, 2019 11:18 pm
You will, however, need to know the dn/dc values of the protein and DNA, and UV extinction coefficients at the measurement wavelength (260 or 280 nm, usually).
The dn/dc values of most proteins and DNA are universal to good approximation - 0.186 and 0.17, respectively. The extinction coefficients can be calculated from the sequence, though it is best to measure them.
The dn/dc values of most proteins and DNA are universal to good approximation - 0.186 and 0.17, respectively. The extinction coefficients can be calculated from the sequence, though it is best to measure them.
3 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: John Guajardo and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: John Guajardo and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science