In combining SEC-MALS with other biophysical techniques, why do you see aggregates in light scattering but not in UV?
-
- Posts: 23
- Joined: Thu Feb 21, 2019 11:18 pm
Like SAXS (small-angle X-ray scattering), light scattering intensity varies with the square of the molecular volume – in other words, it is proportional to the product of molar mass and concentration – whereas UV absorption is only proportional to the concentration. Therefore, light scattering is very sensitive to large aggregates which occur at concentrations too low to detect by UV, and even small amounts will produce a large light-scattering signal.
[answer provided by Kushol Gupta, Ph.D., University of Pennsylvania]
[answer provided by Kushol Gupta, Ph.D., University of Pennsylvania]
2 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science