- Posts: 1
- Joined: Tue Feb 26, 2019 5:24 pm
Detectors for molar mass
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
2 posts
Page 1 of 1
Is MALS preferable over LALLS/RALLS detectors for molar mass determination?
-
- Posts: 23
- Joined: Thu Feb 21, 2019 11:18 pm
Definitely. LALLS (low-angle laser light scattering) uses a single low-angle detector to approximate the scattering at zero angle while avoiding issues of angular dependence of the light scattering, but suffers from excessive noise because any contaminates in the mobile phase or dirt on the optics show up in the low-angle data and compromise the accuracy of the results.
RALLS (right-angle laser light scattering) uses a single detector at 90° scattering where the data are typically much less noisy, but if the molecules are above ~ 10 nm in radius, then one has to correct for the size- and shape-dependent angular asymmetry of light scattering. Hence both RALLS and LALLS are subject to systematic errors.
MALS (multi-angle light scattering) analysis extrapolates from multiple low-noise angles to overcome both the noise and angular asymmetry issues. In addition, MALS can determine the rms radius of the molecule if it is above 10 nm or else deduce that the size is less than 10 nm, both of which are not possible with either RALLS or LALLS.
[answer provided by Steven Trainoff, Ph.D. and Chris Deng, Ph.D., Wyatt Technology]
RALLS (right-angle laser light scattering) uses a single detector at 90° scattering where the data are typically much less noisy, but if the molecules are above ~ 10 nm in radius, then one has to correct for the size- and shape-dependent angular asymmetry of light scattering. Hence both RALLS and LALLS are subject to systematic errors.
MALS (multi-angle light scattering) analysis extrapolates from multiple low-noise angles to overcome both the noise and angular asymmetry issues. In addition, MALS can determine the rms radius of the molecule if it is above 10 nm or else deduce that the size is less than 10 nm, both of which are not possible with either RALLS or LALLS.
[answer provided by Steven Trainoff, Ph.D. and Chris Deng, Ph.D., Wyatt Technology]
2 posts
Page 1 of 1
Who is online
In total there are 3 users online :: 1 registered, 1 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: John Guajardo and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: John Guajardo and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science