Does conformation affect the molecular weight determination by MALS, and why? Do you assume that the protein/polymer must be spherically shaped? If not, how good is the fit with the MALS method?
-
- Posts: 23
- Joined: Thu Feb 21, 2019 11:18 pm
The peptides, proteins or polymers do NOT need to be spherical for their molecular weight to be determined by MALS.
Molar mass is determined from the relationship between light scattering intensity, concentration and scattering angle. Regardless of conformation, the angular dependence of the intensity of scattered light P(θ) always approaches a value of one as the scattering angle θ approaches zero:
In MALS analysis, the data across multiple scattering angles is extrapolated to θ = 0 in order to determine the molecular weight accurately and with no dependence on conformation. For most proteins and smaller polymers, it is even easier: Rg/wavelength(lambda) is so small that extrapolation is not necessary, and P ~ 1 for all angles.
[answer provided by Vincent Hsieh, Ph.D., Wyatt Technology]
Molar mass is determined from the relationship between light scattering intensity, concentration and scattering angle. Regardless of conformation, the angular dependence of the intensity of scattered light P(θ) always approaches a value of one as the scattering angle θ approaches zero:
In MALS analysis, the data across multiple scattering angles is extrapolated to θ = 0 in order to determine the molecular weight accurately and with no dependence on conformation. For most proteins and smaller polymers, it is even easier: Rg/wavelength(lambda) is so small that extrapolation is not necessary, and P ~ 1 for all angles.
[answer provided by Vincent Hsieh, Ph.D., Wyatt Technology]
2 posts
Page 1 of 1
Who is online
In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot], Hollow and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot], Hollow and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science