How much protein do you use in a SEC-MALS experiment?
-
- Posts: 23
- Joined: Thu Feb 21, 2019 11:18 pm
Typically SEC-MALS requires 50-100 µg, though when necessary as little as 10 µg, assuming the protein elutes well off the column. Experiments aimed at testing self-association will often go to relatively high concentration and use more sample.
[answer provided by Ewa Folta-Stogniew, Ph.D., Director, Keck Biophysical Resource, Yale University School of Medicine]
[answer provided by Ewa Folta-Stogniew, Ph.D., Director, Keck Biophysical Resource, Yale University School of Medicine]
2 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google Feedfetcher and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google Feedfetcher and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science