- Posts: 7
- Joined: Fri Aug 21, 2009 4:13 pm
First Background:
We perform SEC measurements routinely and we ran into a problem with a recent sample. The polymer was Poly (N-phenylacrylamide) and the mobile phase was THF at 25 degrees C. We are running the following setup:
Waters 2695 at 0.3 ml/min
Injections of 100 uL
Sample concentration 1 mg/mL
Waters Styragel column bank (4 years old)
Waters 2996 PDA detector
Wyatt MiniDawn light scattering detector
Wyatt OptilabREX differential refractive index
Situation:
The problem is very erratic behavior of the detectors after the "solvent" peaks elutes. In addition the light scattering detector shows a very long tailed peak (even upon terminating data collection) and it took almost all night to return to baseline. Blank THF samples injected after this long tailed polymer show some strange behavior around the solvent peaks and the polymer peaks of standard injections are not as reproducible as they used to be. I think I have adsorbed polymer in my column and I want to know if I can save these columns or not.
Im currently letting mobile phase run through the columns for an extended period of time but if anyone has any suggestions on how to "clean" the columns it would be appreciated.