Hello,

I have a Varian Spectra 55B and i've been analyzing cadmium in water samples, most of the samples are negative for cadmium but there are a few that give me false positive, i'd sent the samples to another laboratory and analyze with ICP and the samples are negative. I was thinking that the problem is in the flame but im not sure...i'd tried the interferences that says in EPA method, tried with KCl solution, filtration,reduce the width of opening window...but all went wrong...what can i do?
control wrote:
...what can i do?

Perhaps find a forum devoted to AAS :-)
i know this is a chromatography forum but i read some posts about AAS, i didnt find an AAS forum, do you know one?
Unfortunately no.
The standard wavelenght is 228.8 nm for cadmium , but did you try also the wavelenght 326.1 nm ( which is less sensitive ) ?

The flame must be oxidizing ( blue ) and a slit width 0.7 nm recommended.

What are the standard concentrations ?
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry