If you get a zero absorbance for a sample, is the concentration therefore, zero?

Because if you create a calibration curve and solve the equation for the line, you don't get a zero concentration.
Depends if you have subtracted your solvent blank and you are in the linear range. Are you talking about UV-vis?
Where can I buy the kit they use in CSI?
It doesn't really matter. All you can ever say about zero response (or response that's in the noise of the detector) is that it's undetected or less than the limit of detection (LOD).
Because if you create a calibration curve and solve the equation for the line, you don't get a zero concentration.
The y-intercept will almost never be *exactly* equal to zero. You need to look at the standard error as well. If the standard error interval does not include zero, then your calibration curve has issues (e.g., a contaminated blank, or outside the linear range, or significant heteroscedasticity in your data).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
OK thanks. I have another query about the equation for a calibration curve. I have run some standards through a UV/VIS spectrophotometer. I have a calibration curve with concentration in mg/L.

I get this equation from the calibration curve:

y = 0.0.7755x - 0.0780

So to calculate the concentration of my sample I have substituted in the absorbance of the standard into y which is 0.0225 and solved for x.

0.0325 = 0.7755x - 0.0780

0.0325 + 0.0780 / 0.7755 = x

I get a concentration of 0.142mg/L


I would also like to calculate concentration by using the formula: c = A/el

As I understand it, e (molar extinction coefficient) would be 0.7755 from the same equation:

0.0325 = 0.7755x - 0.0780

c = 0.0325/0.7755 x 1 = 0.045

I get a concentration of 0.045mg/L

Why do I get a different concentration value? There's something I'm missing.
As I understand it, e (molar extinction coefficient) would be 0.7755 from the same equation:
Several flaws in that statement:

1. That would be true only if the y-intercept were zero.
2. What you're looking at is not *molar* extinction coefficient, but *mass* extinction coefficient (you need to know the molecular weight to convert one to the other). This is a nitpick, because the difference in your value comes from the y-intercept.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
tom jupille wrote:
As I understand it, e (molar extinction coefficient) would be 0.7755 from the same equation:
Several flaws in that statement:

1. That would be true only if the y-intercept were zero.
2. What you're looking at is not *molar* extinction coefficient, but *mass* extinction coefficient (you need to know the molecular weight to convert one to the other). This is a nitpick, because the difference in your value comes from the y-intercept.



So how do I convert mass extinction coefficient to molar extinction coefficient?
By converting mass to moles (i.e., divide mass my molecular weight).

It actually doesn't matter whether you do your calibration in mass or moles, and it's more convenient to use mass. Just be consistent.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
tom jupille wrote:
By converting mass to moles (i.e., divide mass my molecular weight).

It actually doesn't matter whether you do your calibration in mass or moles, and it's more convenient to use mass. Just be consistent.


I'm not sure I understand what the 0.7755 in the equation actually is. Is this a mass e.g. 0.7755mg?
I'm not sure I understand what the 0.7755 in the equation actually is. Is this a mass e.g. 0.7755mg?
No, it's the slope of your calibtation line. If you remember from your first-year algebra class, the slope is "the rise over the run" (i.e., y/x), so the units in this case would be (mg/mL)/AU.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Hi walker8476

Why your absorbance is so small I think 0.0325 is very low. You must use a value of absorbance in the middle of your calibration curve, say 0.3 or 0.4, is this an assay?

Fernando.
11 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry