-
- Posts: 95
- Joined: Fri Nov 30, 2012 5:18 pm
Robustness test for Related Substance
Off-topic conversations and chit-chat.
3 posts
Page 1 of 1
Can I use Change in Detector for Robustness Test? Is it okay if I'm going to use PDA detector first and changing it into 2487 UV/VIS detector? Thanks.
-
- Posts: 82
- Joined: Tue Nov 27, 2012 11:26 am
I would say no. Changing the detector is changing the method. Robustness should only alter the method conditions.
-
- Posts: 37
- Joined: Tue Feb 20, 2007 12:36 pm
jayroseville58 wrote:
Can I use Change in Detector for Robustness Test? Is it okay if I'm going to use PDA detector first and changing it into 2487 UV/VIS detector? Thanks.
Of course you can! This way you will have established method robustness over both UV-detector models. Especially in the future as you might need to transfer the method to another laboratory with PDA/DAD detectors only.
3 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science