Cell Size on UV
Off-topic conversations and chit-chat.
6 posts
Page 1 of 1
What if, there is no available 0.5cm cell size for UV analysis. Can we just adjust the standard and sample concentration indicated on the USP monograph so that we can use the available 1cm cell? If we can, then are we still going to conduct linearity or perform validation instead of verification base from the adjustment we made?
-
- Posts: 405
- Joined: Wed Jul 18, 2012 6:45 am
- Location: Germany
Is this HPLC-UV or photometry using UV?
If photometry, I cannot imagine there are no 0.5 cm cells available (at least here in Germany we have http://www.hellma-analytics.com/startse ... /home.html they can provide every imagineable format).
And HPLC-UV: I don't even know the lightpath of my UV cells, do I have to worry about that?
Jörg
If photometry, I cannot imagine there are no 0.5 cm cells available (at least here in Germany we have http://www.hellma-analytics.com/startse ... /home.html they can provide every imagineable format).
And HPLC-UV: I don't even know the lightpath of my UV cells, do I have to worry about that?
Jörg
-
- Posts: 273
- Joined: Mon Aug 04, 2014 8:13 am
- Location: UK
Hello
Is it UV detector cell or UV spectrophotometer cuvette?
Regards
Tomasz Kubowicz
Is it UV detector cell or UV spectrophotometer cuvette?
Regards
Tomasz Kubowicz
-
- Posts: 95
- Joined: Fri Nov 30, 2012 5:18 pm
kubowicz.tomasz wrote:
Hello
Is it UV detector cell or UV spectrophotometer cuvette?
Regards
Tomasz Kubowicz
UV Spectrophotometer Cuvette
-
- Posts: 2071
- Joined: Tue Aug 17, 2004 7:59 pm
- Location: 38.590967, -90.213236
You would want to dilute everything by an extra factor of 2 so that you should be in the linear range of the method you are using, then calculate normally and report the deviation from the method in the assay documentation. In some labs, you need permission to do this sort of thing in advance.
PS - make sure your 1CM cuvettes are made of the same thing your missing, method specified, 5mm cuvettes are made of...
PS - make sure your 1CM cuvettes are made of the same thing your missing, method specified, 5mm cuvettes are made of...
Thanks,
DR
DR
-
- Posts: 3508
- Joined: Mon Aug 30, 2004 11:43 pm
I think it would be simpler to just buy 0.5 cm cuvette; if not: do what DR posted.
6 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science