Does anyone have perform validation to determinatio for Isocyanate contern (prepolymer) by titration? If yes how you do for accuracy and linearity? Thank you in advance.

Luci
Accuracy for a titration would be (as for HPLC) to spike and assay placebo samples, or to do as standard additions. Precision would be to assay the same sample multiple times. Linearity for many titration assays should be able to be assumed, but you could make varying concentrations and calculate the amount of titrant required.

For example, antiperspirant products are typically assayed for active by titrating the chloride content. And we have documented and validated the procedure with these discussed parameters here. Our QA says that the USP procedure is only for USP raw materials, so must be validated to use on finished products.

It's a living....
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry