BHolmes » Tue Jul 02, 2013 8:40 pm
Our application is multiresidue pesticide analysis on fruits and vegetables (extracted using modified QuEChERS) using LC-MS/MS and GC-MS/MS. A sample sequence for a batch of 21 samples for either instrument is as follows:
2-4 condition injections of reagent blank, matrix blank, low level standard
calibration standards x4 levels (1x, 2x, 5x, & 10x LOD), 1 inj each
1 inj solvent blank (MeOH for LC-MS/MS, none for GC-MS/MS)
1 inj reagent blank*
1 inj matrix blank*
1 inj spiked matrix* (for recovery check)
1 inj standard that matches spike levels in spiked matrix recovery, for comparision only
1 inj sample 1
1 inj sample 2
....
1 inj sample 8
calibration standards x4 levels (1x, 2x, 5x, & 10x LOD), 1 inj each
1 inj solvent blank (MeOH for LC-MS/MS, none for GC-MS/MS)
1 inj sample 9
...
1 inj sample 21
calibration standard x4 levels (1x, 2x, 5x, & 10x LOD), 1 inj each
1 inj reagent blank (MeOH for LC-MS/MS, none for GC-MS/MS)
*extracted using the same modified QuEChERS that used to extract samples
For our purposes we bracket the sequence after every 10-15 inj of non-calibration standard injections to show that we can show calibration integrity throughout the entire run. For LC-MS/MS we use ALL standard injections to generate calibration curve, but for GC-MS/MS only the first set of standard injections are used - don't know why, I'm not the primary chemist for this instrument. For both instruments, we do NOT force through origin (zero) and use simple linear regression analysis to obtain calibration equation in the form of y = mx + b. For LC-MS/MS the curve is weighted as 1/x. For GC-MS/MS the curve is NOT weighted.
Hope this helps! What are your current lab practices for setting up sequence?