Concentration calculation from HPLC

Basic questions from students; resources for projects and reports.

6 posts Page 1 of 1
Hi there,
Can someone help me confirm the calculations below please:

I extracted 120 mg of my sample (79% moisture) into 10 mL solvent. I then injected an aliquot of the extract into the HPLC and got a concentration of 7 mg/L.
41mg = 7mg/L
=0.17mg/mg of sample
% w/w=17.1

Is my calculation of the concentration in the original sample and the weight percent correct?
Thank you
GarryP wrote:
Hi there,
Can someone help me confirm the calculations below please:

I extracted 120 mg of my sample (79% moisture) into 10 mL solvent. I then injected an aliquot of the extract into the HPLC and got a concentration of 7 mg/L.
41mg = 7mg/L
=0.17mg/mg of sample
% w/w=17.1

Is my calculation of the concentration in the original sample and the weight percent correct?
Thank you


Where do the 41 mg come from???

w/w = [concentration of analyte] / [concentration of sample] (both at the same stage of the sample prep e.g. in the vial)

the latter one is
120 mg * (1-0.79) / 0.01 L = 2520 mg/L
So amount is:
7 mg/L / 2520 mg/L = 0.278% w/w (on dried basis)
Thank you for your input. The 41mg was a mistake I made in deducing the dry weight content of the sample.

If I want to simply report the concentration in the sample, would that be 7mg/L or do I need to multiply it by 10 mL (the volume in which the sample was extracted) to give 70 mg/L?
GarryP wrote:
If I want to simply report the concentration in the sample, would that be 7mg/L or do I need to multiply it by 10 mL (the volume in which the sample was extracted) to give 70 mg/L?


No,no,no, why would the concentration change if you just take more volume of it? Does a small beer have a lower alcohol conc than a big one...?
Also the unit would not be correct if you multiply mg/L * L = mg (L cancels out). Even more mg/L * ml = ug, so 70 mg/l would be totaly wrong.

So in 10ml there are 70ug of analyte, those coming from the 120 mg of sample = amount in the sample "as is" 583*10-6 w/w (=583 ppm)

Same is obtained from previous formula, if you don't correct for the moisture.
Something I try to teach my students too:
be strict in using the right terms and check the units

A concentration is always something per volume, so the unit has to be to (something, e.g. g or mol)/L

An amount is mass per mass or unit-peace, so the unit is weight/weight (or pce), e.g g/g or g/tablet

If the unit is just like g, then call it what it is, a mass

further, if not used to it, just write the SI-prefixes as 10^-3 for milli, 10^-6 micro etc, then do the numerics and finaly convert it back to a SI-prefix; e.g mg/l*ml = 10-3g/l*10-3l = 10-6 g = ug

This may help preventing confusion.
7 mg/L is concentration determined by the HPLC

7mg/L is the concentration of the 10ml final solution of your extract.

10ml of extract = 0.01L extract

10ml of extract contains (7mg/L) * 0.01L = 0.07mg of analyte

10ml of extract came from 120mg of sample therefore:

120mg sample contains 0.07mg of analyte. which is 0.07mg/120mg

amount of analyte per unit weight of sample = 0.000583mg/1mg which is 0.0583%

Corrected for moisture 0.0583/(1-0.79) = 0.2778%

Just another way to look at the problem.
The past is there to guide us into the future, not to dwell in.
6 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry