I am having some problems with my Agilent Eclipse Plus C-18 RRHD (2.1x50mm) column, so I am hoping you can help me. I am a student at University of Rijeka, and I am doing my thesis on determination of metallothionein by liquid chromatography (UHPLC). I am currently working on my standard which is comprised of gluthatione dissolved in water and DTNB (Ellman's reagent) dissolved in methanol. For mobile phases I am using phosphate buffer (0,2 M, pH=2,5) and acetonitrile with 0,1% formic acid. A week ago, during my analysis, the pressure suddenly rose to 1000 bar and the analysis was immediately stopped. I checked all the components of UHPLC but the problem seems to be in the column. My assumption was that salts from buffer clogged the column so I tried backflushing it (1h water, then 1h acetonitril- water 50%-50%), but the pressure is still pretty high. I am hoping you can give me some advice as to what else I might do to better the performance of my column.
Phosphate will precipitate when the ACN concentration is above ~50%. Reversing and washing the column with water will help.

Then, try washing the column (at least 10X column volumes) with 100% Methanol, 100% IPA, 100% Hexane and back again.
Hi elizabeta.rus,

You said,

For mobile phases I am using phosphate buffer (0,2 M, pH=2,5) and acetonitrile with 0,1% formic acid.


Is the phosphate concentration Really 0.2 M? If so, perhaps it may be worth some time to see if a lower concentration of phosphate could afford both good peak shape and separation.

I agree with HPLC Chemist's suggestions--warming the water for the backflushing may help to remove salt from the column, as well as using a low flow rate (<100 uL/min) to allow time for the water to dissolve the phosphate salt.

EDIT: This link describes some buffer solubilities:

http://files.pharmtech.com/alfresco_ima ... -97119.pdf
MattM
Thank you HPLC chemist!
Thank you very much mattmullaney!
5 posts Page 1 of 1

Who is online

In total there are 5 users online :: 0 registered, 0 hidden and 5 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 5 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry