Page 1 of 1
HPLC-UV method for nucleotides
Posted: Tue Aug 01, 2017 2:33 pm
by ikik
Hi all,
I need to analyse nucleotides using HPLC-UV. I have a supelco c18-T column. Does anyone have a hplc method I could use?
Re: HPLC-UV method for nucleotides
Posted: Wed Aug 02, 2017 12:27 pm
by HPLC chemist
Re: HPLC-UV method for nucleotides
Posted: Wed Aug 02, 2017 1:14 pm
by ikik
Hi HPLC chemist,
Yes I did and only one of the methods worked but i had baseline problems where the baseline drifted quite significantly affecting my chromatogram.
The method I tried:
Mobile phase A: 0.03M KH2PO4, 10mM TBAHS(tetrabutylammonium hydrogen sulphate) adjusted to ph 6 with NaOH.
Mobile phase B: Mobile phase A:methanol (80:20 v/v)
injection vol: 10 microlitre, flow rate: 0.7 ml/min, column temp: 35oC
When I run it without TBAHS, I have no problems with the baseline. However I need TBAHS to separate the nucleotides.
Re: HPLC-UV method for nucleotides
Posted: Fri Aug 04, 2017 6:02 pm
by HPLC chemist
Sounds like your 'buffering/ion pairing' is not stable for your sample matrix. Increase the concentration of the TBAHS to 25 mM.
Re: HPLC-UV method for nucleotides
Posted: Fri Aug 04, 2017 6:04 pm
by HPLC chemist
Also increase the concentration of the Phosphate buffer to 50 mM (not 0.03 mM).