I'm trying to sort out how best to quantitate ethylene in air. Most readily available is GC/FID; the crop (carnations) is sensitive to ethylene in single-digit ppb levels.

I don't suppose FID does well with ethylene at those concentrations, does it? If not, what're my options in terms of sample prep etc.?
Single-digit ppb might be tough. Ethylene is not easy to concentrate. You might be able to beat FID detection limits with pulsed-discharge helium-ionization detector. Yours is a tough problem.
osp001,

Challenging. Not a good candidate for HID since air is the matrix and it would have to be very well resolved. Besides, ppb with HID can be very challenging.

Should be a reasonable responder on argon ionization detector (IP = 10.5) but I have not had great luck with it.

Not going to concentrate very will with a very low breakthrough volume.

If you want to try AID let me know (aicmm at flash (dot) net.) Otherwise I would suggest RGD but might not like the chromatography very much.

Best regards,

AICMM
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry