RI vs ELS detectors for detection of polysaccharides

Basic questions from students; resources for projects and reports.

3 posts Page 1 of 1
I need to characterize dextrans and their non-UV active derivatives, and most of the literature I see uses refractive index detector. Is it possible to use evaporative light scattering? I always thought that it is more sensitive, stable and robust. What would you say?
Is it possible to use evaporative light scattering?
Yes: https://duckduckgo.com/?q=dextran+elsd&t=ffsb

ELSD pros:
- probably more sensitive
- less of an issue with temperature control
- compatible with gradients
ELSD cons:
- more complex (need for N2 source)
- harder to set up (more parameters to fuss with)
- mobile phase buffers must be volatile
- limited linear range

RI pros and cons pretty much the inverse.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thanks. This is very helpful.