Inconsistent rentention time HELP. URGENT~!!!

Basic questions from students; resources for projects and reports.

18 posts Page 2 of 2
If you inject the same standard repeatedly do you still see the retention time drift?

Your differences in RT seem pretty negligible to me. Especially over such a large concentration range. But then I don't work in a regulated environment or with nice clean samples.
richiekichi wrote:
If you inject the same standard repeatedly do you still see the retention time drift?

Your differences in RT seem pretty negligible to me. Especially over such a large concentration range. But then I don't work in a regulated environment or with nice clean samples.


Thanks for your opinion, richiekichi.

I lower my concentration and its range, so the difference RT wonldn't be so large.

Problem solved :)
Apart from the issues already mentioned, decreasing retention time with increasing concentration may be due to column overload.
18 posts Page 2 of 2

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry