I'm having trouble with this question:

The capacity factor for nicotine separated on a C8 non-polar stationary phase is shown
below.


Capacity Factors for Nicotine at three Different pH values

pH3 pH5 pH7

0.00 0.13 3.11

Why does the capacity factor increase as the pH increases?

The pka of nicotine is 3.15, so at a pH above 3.15, nicotine will become deprotonated and therefore have a negative charge which makes it more attracted to the polar mobile phase which means it moves faster through the column and therefore the retention time will decrease and so capacity factor should decrease.

What is wrong with my thinking?
This may help: is nicotine an acid or a base?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
tom jupille wrote:
This may help: is nicotine an acid or a base?


Base
walker8476 wrote:
tom jupille wrote:
This may help: is nicotine an acid or a base?


Base


OK I see where you're going. So nicotine is a base, so as you increase the pH, the nicotine molecules that are un-protonated will remain un-protonated and the nicotine molecules that are protonated will become un-protonated and therefore become neutral and therefore less polar, which means they will be more attracted to the non-polar stationary phase which increases the retention time and therefore the capacity factor increases also.

Is that better, or is this just as a bad as my first attempt?
You got it! :D
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
tom jupille wrote:
You got it! :D


Can't believe that was right. Thanks for your help!
6 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Majestic-12 [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry