Hello dear colleagues, we have a RID-10a detector, that is working very well. But there are some features that I don't understand. In the recorder range feature it goes from 0.01 to 500 uRIU/FS(Full Scale). When is better to use 0.01 and when 500? How can we interpret this 2 values?. the meaning of full scale (FS)?.

Thanks a lot for your kind cooperation.
"Full scale" is just what it sounds like: the value in RI units at the top of the chart.

The setting should match the height of the largest peak of interest. So for example if your largest peak has a maximum height of 8 x 10^-5 RI units (i.e. 80 uRIU), then the scale should be set at 100 uRIU/FS.

If the maximum height exceeds full scale, the peak will be "flat-topped" and quantitation will be incorrect.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
tom jupille wrote:
"Full scale" is just what it sounds like: the value in RI units at the top of the chart.

The setting should match the height of the largest peak of interest. So for example if your largest peak has a maximum height of 8 x 10^-5 RI units (i.e. 80 uRIU), then the scale should be set at 100 uRIU/FS.

If the maximum height exceeds full scale, the peak will be "flat-topped" and quantitation will be incorrect.


Tom Jupille, thanks alot for your reply. The example was very clear to solve my doubt. Thank you very much..!!
3 posts Page 1 of 1

Who is online

In total there are 8 users online :: 0 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry