Hi,

I have read that column ovens are important for sugar analysis. I always assumed this was to promote better and consistant column separation however I recently read that when using RI detectors (commonly used for sugar detection) that it is important to use a column oven to maintain a constant temperature and avoid baseline shift. I know that RI is temperature dependant but surely any eluant leaving the column oven will cool before it passes through the detector. Shouldn't the cells in the detector be heated to maintain a constant temperature of samples passing through? Thanks in advance for clarifying this. :lol:
0.01" id tubing has about 0.5 microliter of volume per centimeter of length. Assuming you have 20 cm of tubing between column and detector, that works out to 10 microliters. At 1 mL/min, that's a residence time of 0.010 x 60 sec = 0.6 seconds. Not much time to cool down.

And yes, many RI detectors incorporate heat exchanger tubing to damp out temperature fluctuations.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thanx for the reply!
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry